I have an EUW-24 VTVM here which was given to me by the U. of I. surplus.

It has some problems, and I would like to fix those and use the instrument.

The EUW-24 was a factory-built version of a kit VTVM, but I am not sure 
which kit that was.

In any case, I need a manual for my unit or for the kit one that is identical 
with it.

I have a schematic and the calibration procedure for the EUW-24, and have 
built a proper probe for the VTVM.

The problems are: 

1) when switching my probe to the DC position (with the 1 meg resistor in 
line), when attempting to measure a 1.5 V drycell, the reading drops to zero. 
If I have the probe set to AC/OHMs, the reading of a 1.5 V dry cell is close to 
accurate. I have used an external digital VOM to check the probe, and with 
the switch in the DC position, I read almost exactly 1 megohm. With the 
switch in the AC/OHMs position, I read something like 0.5 ohms.

2) At OHMs settings of X100,000 and above (the last three), I cannot set the 
meter to read infinity with the probe open. I can set the probe to read infinity 
for all but the top three settings. I can set the meter to zero with the probe 
shorted to ground in all settings. 

I thought that perhaps the coax in the probe was leaky, so I unplugged the 
probe. There was no difference.

I am hoping that a manual will help me figure it out.

Ken W7EKB