Hi Dick
Are you measuring/assessing the vertical sensitivity using a signal
generator or a rig?
The resonant frequency of the tuned circuit using inductor L1 is dependant
on the capacitance of the coax cable connecting the SB610 to the rig plus
any coax inside that rig and the small coupling capacitor to the receiver
last IF amplifier. If I recall correctly Heathkit recommended the use of high
impedance (93ohms) cable because it has a lower capacitance per unit
length than conventional 50 ohm type. You need to get that tuned circuit to
show resonance at the correct frequency in order to get the correct
sensitivity.
Check the resonance with a GDO to see where it is currently located.
If you connect a signal generator to the vertical input the resonance of L1
will not be apparent because of loading by the generator output impedance,
typically 50 ohms.